Last edited by ashiekh; 03-26-2019 at 09:52 PM.
for job use differential probe. or some scope have built in isolate input like tektronix TPS series or some hioki, picoscope scope. really expensive than normal scope ( with same features)
just a hobby. cut off ground cable in your scope power cord. always use 10x and up passive probe. cheapest way but dangerous for young player
Or just use the dual channel method and stay safe.
NEVER cut the ground wire.
Last edited by ashiekh; 03-27-2019 at 09:21 AM.
It's not safe.
Show me a general scope with anything but a CAT-I rating.
You are measuring a circuit connected to mains here.
The scope needs at the very minimum a CAT-II rating for such measurements.
Either use the proper equipment (a differential probe) or just don't do it.
There is no point for the average person to connect a scope up to mains and kill himself in the process!
"The one who says it cannot be done should never interrupt the one who is doing it."
I thought most scopes were good to 300V; the one I have access to is good to 400V
I stand by what I have written in my previous post and see no reason to change any recommendations within.
A standard oscilloscope is designed for CAT-I environments, the primary side of a PSU is a CAT-II environment.
"The one who says it cannot be done should never interrupt the one who is doing it."
So you are saying that most 'scopes shouldn't be used for mains voltages?
Now it is true that 240V AC is 340V peak, but in the States we are half this.
Last edited by ashiekh; 03-30-2019 at 05:31 PM.
Many thanks, this has been an education; the reason why I visit this forum.
I would imagine most oscilloscopes are CAT I rated exactly because of the grounding issue, which would make them unsuited for direct mains measurements; the dual channel trick removes this restriction.
Last edited by ashiekh; 04-01-2019 at 02:08 AM.
Isolation transformers are a very useful tool.
Scopemeters are very nice if pricey.
Capture and data connectivity is common now, but it's been available for many years.
1) It never hurts to probe for voltages before working bare hand
2) De-energize whenever possible
3) If in doubt, one hand in the pocket, wear isolation tested footwear
4) RCD protected circuits can save lives
5) Always be aware of arc flash risks