    Default Brand Power Meter Accuracy

    So recently we had some issues with our Extech power meter in the lab. A couple of units failed during testing while connected to it, including one that gave a dead short, and the unit hasn't worked right since, even after RMA. So I proposed we use a Brand 20-1850 instead, at least until we could get a higher end meter.

    We picked up the Brand meter and I was initially quite pleased. However, when I tested the first unit with it, I found that I was getting efficiency results about 5% above expected values! The unit was reading 5% low.

    Well I looked up the calibration guide, then constructed a precision resistive load using 1% low temp co 50W 500 ohm resistors. Using that and my Fluke meter to measure the AC voltage, I could get accuracy to 1% in the 50-120W range. However, outside of that range, again, the reading is off by increasing amounts as I go up in wattage. Overall, we were 1-3% out compared to the Extech and the results from our EU and Taipei labs.



    I called up Brand Electronics and spoke to Ethan, who seems to be the engineer and sole proprietor. Apparently the Brand meters are only 2-3% accurate in practice. He offers calibration service to get them to 1% accurate at a specified load, and then provide a cal curve for higher and lower loads, but the unit itself cannot be calibrated tighter than that.

    However, he does offer a higher-precision version, which has a more robust analog frontend and uses a 24-bit ADC, instead of the uC's native ADC, which he can cal to 0.5%, or can be caled by a NIST lab to 0.1%, with low drift. This is a "semi-custom" part according to Ethan, which is why it isn't listed on the website, but he offers it for $350 on request.




    I thought I'd just share this info for anyone reviewing units using a Brand power meter. Keep in mind that these have a distinct error curve outside of their calibrated wattage. The "semi-custom" version may be a better solution.

  2. The Following 2 Users Say Thank You to CM Phaedrus For This Useful Post:

    Philipus II (03-30-2014), tinpoint (03-26-2014)

  3. #2
    Join Date
    Oct 2006
    Location
    jonnyGURU forums, of course!
    Posts
    16,084
    Thanks Thanks Given 
    540
    Thanks Thanks Received 
    274
    Thanked in
    198 Posts

    Default

    This is good info, but I want to put it in the correct sub-category.

  4. #3
    Join Date
    Jul 2013
    Location
    Chino Hills, CA
    Posts
    282
    Thanks Thanks Given 
    16
    Thanks Thanks Received 
    1
    Thanked in
    1 Post

    Default

    No problem, sorry for the wrong section!


    Based on my discussion with Ethan, the sources of inaccuracy in the 20-1850 are these:
    - Accuracy and resolution of the main uC ADC (16-bits I believe)
    - Anti-aliasing to compensate for low resolution
    - Current noise below ~200mA in sensing circuit
    - Component accuracy rating, temperature co-efficient, and drift
    - Power supply noise (he's using a regular 5V LDO and bypass caps)

    The "semi-custom" version addresses the first three issues and improves the fourth, thereby achieving much greater accuracy.

Similar Threads

  1. Consumer grade power meters accuracy
    By quest for silence in forum Testing Methodology Discussion
    Replies: 11
    Last Post: 06-24-2015, 08:19 PM
  2. Integrated Power Meter
    By Mach in forum PC Power Supply Discussion
    Replies: 7
    Last Post: 01-17-2011, 11:39 PM
  3. My new Power Meter :D
    By crmaris in forum Testing Methodology Discussion
    Replies: 4
    Last Post: 12-21-2010, 09:07 PM
  4. Power Meter with accuracy above 700W
    By SKYMTL in forum Electronic Component Discussion
    Replies: 30
    Last Post: 06-08-2008, 07:37 AM

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •